Hi Peter,
I was cautious to post details due to anonymity. But:
Column = HP-5, 30m x 0.53mm, 1.5 um film
Control = Head Pressure @ 40 kPA
Split = 1:5
Inj Vol = 1 uL
INJ = 270C
DET=FID@280C
Oven=130C, isothermal
Elution order = IPA, THF, DMSO
I've attached an example chromatogram below as well.
We're using DMSO as diluent due to the nature of the sample matrix and this is a method that was validated and transferred to us from another lab so we have little latitude to change conditions or solution preps. Unfortunately I'm unable to contact the transferring lab for answers due to reasons.
My confusion is that the splitting is intermittent and not across all injections and is seen randomly in both standards and samples.
Hopefully this will give you more information to go on.
Thanks,
GL6