Hi
Why am I not able to detect Hydrogen using HaySepC and molecular sieve 5A columns with TCD detector, Helium as carrier gas,
Injection temperature = 200C
TCD Temp = 200C

thanks
Suni
You did not mention the concentration at which you are trying to detect and the amount of gas you are injecting.

There are a lot of reasons why you are not detecting H2.

Detector problems

Leak problems

Injection problems

The low and NEGATIVE response H2 has using He as a carrier on a TCD detector

I hope this will give you enough information to troubleshoot your analysis problem.

Good luck,

Rod
Thanks, Rod.

I have checked the leak,etc..

I tried with 5% Hydrogen in Nitrogen.
The sample flow rate in GC is 15ml/min.

Also using 10 port gas sampling valve.
thanks for the immediate reply....

regards
Suni

H2

You did not state what column you are using. A poor quality column or a poor installation will make H2 difficult to see at only 5% using Helium carrier.

Try using Nitrogen carrier for improved sensitivity.

good luck,

Rod
Two columns - HaySepC and molecular sieve 5A

thanks
Suni
5 posts Page 1 of 1

Who is online

In total there are 49 users online :: 1 registered, 0 hidden and 48 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 48 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry