High pH matrix using LC-MS ESI positive mode

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Need a bit of help please. If I'm extracting analytes and my matrix is so basic that my pH is ~10, when I run the sample, will all the base overwhelm my acidic eluent such that no protonated ions will be formed upon ESI? I'm not seeing any analytes reported in a sample of mine I submitted. I don't have control over the instrument, and can't directly talk to the analyst (touchy reasons). I'm working on a theory that the extract should have been brought to neutrality before running. Any ideas out there?

Thanks,

MR

It depends... on your analytes and the base that was used to have the pH brought at 10.

There are analytes that has been shown to have the same electrospray response at all the pH spectrum (i.e. acidic, neutral and basic).

What is the base used to bring the pH at 10? If it is something like NH3 it's OK. If you use an organic base with high proton/affinity-gas bacisity then you might have significant suppression effects...

As a rule of thump, I always try to bring the pH to acidic -with formic acid/acetic acid- when I am working in positive ion mode (assuming pH stability of the analyte).

Conclusively, yes, I would give it another try. Maybe you even need to do some chromatography to elute your compound in a matrix free environment in order to see it...

Talking to the analyst would also help :wink:

Besides an MS issue, there could be a chromatography issue as well. Assume that your analytes are anions at pH 10, and that these anions are unretained, a high concentration of something like ammonia in your sample could make them elute as unretained peaks. This is roughly the same as injecting a large volume of DMSO.

It would depend on the volume of dissolved analyte you are injecting and also the strength of the % aqueous solvent used in the reversed phase separation.
Sailor

Hmmm,

I made the assumption that there is no chromatography involved after the extraction maybe MR can clarify this ...
5 posts Page 1 of 1

Who is online

In total there are 242 users online :: 1 registered, 0 hidden and 241 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm

Users browsing this forum: Ahrefs [Bot] and 241 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry