How much time was there between replacing the interface ferrule (= venting the MS) and this autotune? Did you follow the instrument guidelines of doing this operation?
I've been to about 6 hours and then proceed tune
And I haven't do this step:
“… Set a minimum velocity of 30 cm/s, or as recommended by the column manufacturer. Allow the carrier gas to flow through the column at room temperature for 15 to 30 minutes to remove air. “
Also new graphite/vespel ferrules will shrink when heated, so after about a day of being installed on the heated interface you will need to retighten the nut about 1/8-1/4 turn to remain sealed. One trick you can use to prevent the ferrule from leaking is to keep the ferrule in the GC oven in a beaker and the temperature cycles will pre-shrink the ferrules so they are ready to go when you need to change one.
Thanks for your trick. I will be applying.
I did like your suggestions,
I opened the analyser chamber and saw column inserted into the interface too long.

and i clipped.
And the results are as follows:
http://i.imgur.com/G9gjw8I.jpg
But, when I scan organochlorine pesticide standard, have 2 problem:
1, mass assignments are incorrect
example: 181 -> 180.8; 165 -> 165.1; 355 -> 354.8 …
2, There are a lot of ghost peaks
What could be the cause? how to troubleshoot? or incorrect tune file (inappropriate parameters)?
Hope to get help. Thanks a lot.