According to Beers' law, the intensity of output light should be exponential relation to concentration. But if this is the case, when we integrate to get the peak area, it is not linear with injection amount. Is there any device in dector transfer the signal of light linear to concentration so that the peak area is linear to injection amount?
-
- Posts: 64
- Joined: Tue Oct 04, 2005 12:26 am
I do not understand exactly , your post. Take any solution of your compound injct 5,10,15,20 micro lit. and find areas for all peaks.Areas must be in linear with the injection amounts.
-
- Posts: 134
- Joined: Sat May 28, 2005 3:34 pm
According to Beers' law, the intensity of output light should be exponential relation to concentration. But if this is the case, when we integrate to get the peak area, it is not linear with injection amount. Is there any device in dector transfer the signal of light linear to concentration so that the peak area is linear to injection amount?
Just give an another try to whatever you have been doing. You will get a linear relation between area vs concentration (Beer's law).
Regards,
Amaryl.
Try again
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
The Beer-Lambert law states that the ratio (given as %, called transmittance, T) of transmitted light (I, or light intensity hitting the sensor) to incident light (Io, or light intensity hitting the sample) relates to the concentration of the light absorbing substance in the cell via a ln function:
-lnI/Io = -lnT = A = abc
where A is the absorbance put out by a photometer (HPLC detector),
a is the absorption coefficient or molar absorptivity, b is the distance the light travels through the solution, c is the concentration of light absorbing substance in this solution. Now, the relationship of c to A is of course linear, that´s all that one has to worry about, normally.
(This relationship can also be nonlinear if, as an example, the substance associates at higher concentrations).
-lnI/Io = -lnT = A = abc
where A is the absorbance put out by a photometer (HPLC detector),
a is the absorption coefficient or molar absorptivity, b is the distance the light travels through the solution, c is the concentration of light absorbing substance in this solution. Now, the relationship of c to A is of course linear, that´s all that one has to worry about, normally.
(This relationship can also be nonlinear if, as an example, the substance associates at higher concentrations).
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
if you are not getting a linear response for changed concentration then 4 things might be going on:
1. your injection volume is too low. (lower physical limitation of detector)
2. your injection volume is too high. (higher physical limit of detector)
3. your compound is doing something else another then just absorbing the light. maybe it has fluorescence, phosporous capacities?
4. your mobile phase is the one "reacting"
any other ideas, any one?
1. your injection volume is too low. (lower physical limitation of detector)
2. your injection volume is too high. (higher physical limit of detector)
3. your compound is doing something else another then just absorbing the light. maybe it has fluorescence, phosporous capacities?
4. your mobile phase is the one "reacting"
any other ideas, any one?
-
- Posts: 18
- Joined: Thu Jan 19, 2006 3:14 am
The injection volume is 10uL
I used 5 different concentrations: 0.11104g/L 0.22208g/L 0.37013g/L
0.5552g/L 1.1104g/L Biphenyl in MeOH.
I changed the uv detector sensitivity( from 0.01 to 0.1) all other setting remained the same. After that a more linear result was gotten (R2=0.97) Is it only the problem of UV detector setting>>??
I used 5 different concentrations: 0.11104g/L 0.22208g/L 0.37013g/L
0.5552g/L 1.1104g/L Biphenyl in MeOH.
I changed the uv detector sensitivity( from 0.01 to 0.1) all other setting remained the same. After that a more linear result was gotten (R2=0.97) Is it only the problem of UV detector setting>>??
6 posts
Page 1 of 1
Who is online
In total there are 194 users online :: 1 registered, 0 hidden and 193 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm
Users browsing this forum: Semrush [Bot] and 193 guests
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm
Users browsing this forum: Semrush [Bot] and 193 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science