The info I can supply is that 1st of all these are negative peaks turned upright by using the negative polarity function in the detector.
2nd the mobile phase in a nitric acid solution, which has greater absorbance than the samples (that's how the negative peaks arise)
3rd the samples are in water only (not nitric) that is probably the initial peak up to 200mV seen just after the injection peak.
The starnge thing is that that sometimes part or all of this disappears.
the first and last chromatograms were fine. Both the 'water' and standard peaks are of the expected areas.
Standard A is from one vial, Standard B is a separate preparation but injections are all from the same vial.
Any ideas appreciated. Thanks for reading.
Standard A run 1 (good)

Standard B run 1 (bad)

Standard B run 2 (bad)

Standard B run 3 (bad)

Standard B run 4 (good)