What I typically run for the SST is 8 injections from the same vial same volume 5ul every injection. What I noticed was the first 6 injections are typically fine, it was the last 2 injections that have a significant drop in peak area resulting in very poor RSD% or CV%
I can't afford to let the CV% go above 5% for any of the analytes as I am running clinical samples.
This assay runs on isocratic mode for 8 mins in each run. We are quantifying 4 analytes including internal standard for each run.
I'm using an AB Sciex Qtrap 5500 that is linked to an agilent 2960 LC system.
What I have tried so far is making fresh standard, fresh mobile phase and also changed the guard and analytical column but still the problem persisted. Have checked with the field specialist who insists that it must be my mobile phase or column. An engineer has came down previously and assured me that the MS is fine. The retention time for all the analytes are also pretty constant, they do elute out at the correct retention times.
The strange thing is sometimes 3 of the analytes the peak areas will be fine while one the last 2 peak areas are severely low, other times 3 analytes can have low peak areas in the last 2 injections while the other analyte is fine for all injections. This is the inconsistency that I am really perplexed about.
Have also flushed the system many times with ACN and water. Also the peak areas do bounce back to normal whenever I inject blanks or re equilibrate the system each day. The peak area does not decrease day by day but rather just the last 2 injections