First, precisely what was done with the pH? Up? Down? What with?
Second, you might try switching from MeOH to acetonitrile or other solvent as your organic because that will alter your selectivity. Also, if your column has active silanols (I have no experience with the column you're using), triethylamine might help if your counter ion is not ammonium or if you switch to a phosphate or other buffer system, but be careful not to raise the pH of the MP beyond what your column can tolerate.
If nothing else, a longer column or a smaller particle size would improve things.
Another option might be to see what effect an ion-pairing agent might have on selectivity. Be VERY careful if you choose to try this route, however, as things can get messy quickly. I was able to do this to good effect once on a reversed phase assay for ascorbic acid in biological media. I didn't have much flexibility with the mobile phase from the standpoint of organic solvent selection because the MP was 100% aqueous, yet I had an issue with a co-eluter arising from the media. The IP agent (alkyl sulfonate) didn't affect the ascorbic acid, but DID move the co-eluter away quite nicely and the assay worked well. If you choose to try this, only do it with an isocratic method, dedicate your column, leave lots of time for equilibration, and be sure to rinse the IP out of your system after you're done with the analysis. As I said, this can work, but should be tried as a last resort.
You may be able to try something on the sample prep side, too. SPE may be a good route as theophylline can be ionized at high pH and there are SPE supports that can work at very high pH's (Waters' Oasis line comes to mind), so you have flexibility there. I might try SPE at a variety of pH's on this sample to see what happens. On a RP-SPE, theophylline would likely be retained at a pH below 8.00 and eluted at 10 or so (pKa = 8.77). This would require validation, however and could actually end up being a fair amount of work. I'd be inclined try to improve your separation first.
Hope this helps or at least gives some fodder for comment!