As the volume of sample you inject increases, then you will get larger peak areas and signal/noise ratios. Thus large sample loops may be essential in some trace analytical applications.
However, to be balanced against this is the effect of volume overload. The volume of sample injected must be small in comparison with the volume of mobile which contains your peak; thus the volume injected is more critical for peaks of low retention compared with peaks of high retention, as Kostas has suggested. As an approximation, a completely filled 20ul loop is probably about the maximum volume I would want to use to avoid seeing too much of this effects. It is not too difficult to calculate the consequences of sample volume on column efficiency for peaks of various k, but it might be easier just to do the experiments and see how your separation is affected by sample volume.
The situation with an autosampler is different. Here the loop is partially filled. When the valve is switched the loop is usually backflushed on to the column, which reduces sample mixing in the loop during the injection process. Thus, 20ul injected from a partially filled 100ul loop might give you very similar results to 20ul injected from a totally filled 20ul loop. I presume you can inject whatever volume you want with your autosampler, partially filling the loop with say 5, 10, 20, 100ul. If you do this you will observe degradation of the efficiencies of the lowest retained peaks, even if you are using quite a large standard bore column (e.g. 4.6mm). The effects will be worse as the column diameter decreases.
If you are doing gradient analyis, then you can inject much larger volumes without any degradation of separation. This is because the sample can be concentrated on the column during the initial stages of the gradient.