I recently started using an old Varian GC 3380 system with the aim of studying migration from recycled paperboard for my PhD research.
I am in the process of developing a detection method for a solution of surrogate compounds. However, i keep struggling with severe reproducibility issues.
Separation and peak shape all seem OK, but the relative standard deviation of repeated injections of the same sample goes up to alarming numbers.
It seems that the deviation of polar compounds is a bit higher than that of apolar compounds.
I have tried almost everything but can't seem to get it right.
Apparatus: Varian GC 3380 with a 1079 injector (single-tapered liner).
Temperature program: hold 0.5min at 100°C, 25°C/min, hold 6min at 280°C
Injection of 1 µl at 10µl/s
Injector at 340°C
Solvent: tert-butyl methyl ether
Split ratio 1/40
Below are some captures of chromatograms.
Full chromatograms:
http://picpaste.com/ales-wNNMvV8K.png
Solvent peaks:
http://picpaste.com/solvent1-4YjNpuiN.png
Crop of solvent peak:
http://picpaste.com/solvent2-VGh86fRI.png
Internal standard peak:
http://picpaste.com/piek1-TdcZo5Wi.png
Analyte peaks:
http://picpaste.com/pieken-vbxufUl0.png
I am rather new to chromatography and completely new to this forum, so feel free to ask for further info. Thank you so much for any suggestions.