						I am trying to do my monthly tune for the Agilent 5973 direct intertion probe. The percentage for 219 is around 20% when its supposed to be 40-85% I did an air and water test, and the nitrogen (m/z 28) was at 111% relative to the m/z 69. How do I go about fixing this
					
									- 
								
 
- Posts: 26
- Joined: Thu Apr 28, 2016 11:45 pm
						Hi
You try tighten the column nut which connect to msd interface
									You try tighten the column nut which connect to msd interface
- 
								
 
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
						Hello
Regards
Tomasz Kubowicz
									There is no column. It is direct insertion probeHi
You try tighten the column nut which connect to msd interface
Regards
Tomasz Kubowicz
- 
								
 
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
						Hello
First check seal between DIP and MSD - it is first suspected place for leak.
Regards
Tomasz Kubowicz
									It looks you have massive leak.I am trying to do my monthly tune for the Agilent 5973 direct intertion probe. The percentage for 219 is around 20% when its supposed to be 40-85% I did an air and water test, and the nitrogen (m/z 28) was at 111% relative to the m/z 69. How do I go about fixing this
First check seal between DIP and MSD - it is first suspected place for leak.
Regards
Tomasz Kubowicz
- 
								
 
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Is the abundance of 219 at the same level as the previous tune at similar EMV settings?I am trying to do my monthly tune for the Agilent 5973 direct intertion probe. The percentage for 219 is around 20% when its supposed to be 40-85% I did an air and water test, and the nitrogen (m/z 28) was at 111% relative to the m/z 69. How do I go about fixing this
If not, it is possible that you are almost out of PFTBA which would make any air present look elevated. If the PFTBA is ok, then it would most likely be a seal around the probe unless you have vented the instrument since the last check. If you haven't vented then you shouldn't have problems with the seal around the side plate.
The past is there to guide us into the future, not to dwell in.
				
			5 posts
							Page 1 of 1
					
	Who is online
				In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science














 
																							