5973 autotune will not pass

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I have changed the source 3 times I just changed the horn everything looks great except the relative abundance of the 219 is less than 35% it used to run around 77% and now I hardly get 25%

This is on autotune mode
standard spectro autotune gets around 44%

any thoughts or ideas im willing to try anything

im using a .15 20m .15 film coulmn and tuning under 50psi head pressure

my foreline runs about 34torr with that setting
2 shimadzu 2010s
3 shimadzu 2010 Plus
1 5973
2 MPX API 4000

Hi!
First, reset your autotune value and make a autotune again.
If not o.k, do a tune evaluation (Msdchemstation, Qualify, Tune evaluation)
What is the reppeler value??How many peak in your tune report?What is your EM voltage??

If reppeler value is 15 to 30 and peak number less than 200, that's o.k.If not, clean your source again.


Do you clean your source part with alumina or with abrasive paper???
2 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry