1. We run system suitability first before proceeding with any validation run and in the result, the RT of cefuroxime is repeatable with RSD of less than 2.
2. During the validation run, we noticed a change in RT for both cefuroxime and IS sulfamethoxazole.
3. RT changed from 6.2 to 6.6 minutes for Cefuroxime and 7.5 to 8.0 minutes for sulfamethoxazole.
M.P.: MeOH/ 0.005 M KH2PO4 buffer (pH 3.6), 30:70
Sample treament: Analytes are in plasma, precipitated with Acetonitrile and back-extracted with DCM/Hexane (1:1)
Stock and Working Solutions are dissolved in 0.01M KH2PO4 buffer, pH 6.5
Is this change in RT significant? Is there any part in the sample preparation or Mobile Phase that can influence the change in RT?
I will appreciate any feedback!
Thank you.
Ghie Malig