in-situ Photo degradation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Has anybody come across of photodegradation in DAD cell (Agilent 1290 DAD)?
basically we see (by MS) potential photo degradant under the main component when we go to MS via DAD detector. If we bypass DAD- the extra component is not detected

I was just wondering how common occurence is this
Probably more common than anyone realizes, because:
a. the DAD hits the sample with broad-spectrum UV (unlike a variable-wavelength detector where only the detection wavelength is used. But
b. nobody looks for it, and
c. it presumably happens the same way with both calibrators and samples, and so any quantitative effects usually cancel out.

That said, I heard about a situation where a monomer photopolymerized in the cell (the chromatography had separated it from the inhibitor). My recellection is that the cell had to be replaced.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry