				GC  OFFLINE PROBLEM
				Posted: Sat Apr 23, 2016 9:20 am
				by y.ALKHADIRI
				Hi
i'm a new chromatographer and new in the forum .
i'am having a problem with the GC i work with . everytime i try to launch offline it ain't open and   i get an erreur message says the location of file does not existe .
why that always happen ? how can i deal with that ? and what file and what location ? 
thank you
			 
			
					
				Re: GC  OFFLINE PROBLEM
				Posted: Sat Apr 23, 2016 5:51 pm
				by dblux_
				There are several different chromatography softwares and different GCs. They behave differently. That's why you have to be more precise if you want somebody to try to help you.
			 
			
					
				Re: GC  OFFLINE PROBLEM
				Posted: Mon Apr 25, 2016 9:35 am
				by y.ALKHADIRI
				i work with agilent technology's GC. 7890B
			 
			
					
				Re: GC  OFFLINE PROBLEM
				Posted: Mon Apr 25, 2016 10:28 am
				by tkubowicz
				Hello
I think the problem you have is with .ini file (it is file that is created during first instalation). If you don't have a lot of experience, I'd recommend:
1. put Chemstation software to CD/DVD drive
2. You should have window with 3 options: repair, modify, remove
3.Choose repair and system will load all files again (overwrite)
Of course backup will be done automatically however you can safe your data before (safe option is to save Chem32 folder).
The procedure I described is the simplest one as you don't have experience with Agilent products. If you have some knowledge about registers and core files you can try to check them.
Regards
Tomasz Kubowicz