I am working with the reversed phase HPLC and gradient method. My hplc pump was running fine before. It has done approx600samples but now sometimes there is no flow from the pump but the pump is on and the light glows.

Do you have any answer why this si happening

thanks
Obvious things to check:

1) Is mobile phase reaching the pump? available? lines crimped or blocked? inlet filter dirty?

2) Is the pump primed with working check valves?

Check what Raven suggested:

1. If pressure exceeding too much; may be blocked line somewhere.
2. If no pressure build up at all; Pump may not be primed. Suck the mobile using the syringe provided.
3. if pressure fluctuating; may be some air entraped in-line.


Most apparent cause to your problem is Pump not primed.
Jitender Madan
Division of Pharmaceutics
Central Drug Research Institute
Lucknow, India
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], John Guajardo and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry