YuYing » Mon Apr 18, 2016 3:30 pm
Hi Peter and Alex,
Thank you for the replies!
Yes, it was the end where gas flowed into as this method was recommended by the GERSTEL Application Chemist. I also took reference from the following literature: P. Werkhoff; W. Bretschneider J. Chromatogr.,1987, 405, 99-106. Could you explain further how does higher trapping temperatures lead to higher recoveries? I am interpreting as the analytes would have to migrate through Tenax all the way at all temperatures, regardless if it is at -50 or -100. I am aware of the normal practice, but I did it the other way because I wanted consistency in each injection by getting the instrument to inject into the Tenax instead of me doing it manually which could have human errors arising.
I was using Tenax-TA in the TDU but the baffled empty glass liner in the CIS. So what happened was that I selected the liquid injection option on the software and the multipurpose sampler gripped the Tenax tube, placed it in the TDU and the solution was injected into the Tenax. The initial temperature of the TDU was held at 25 degrees for 1.5 minutes before I ramped it up to 250 degrees and held it for 10 minutes.
The chemical solution contained a mixture of alcohols, esters, aldehydes, ketones and it was dissolved in Methanol as Tenax has low affinity for methanol. The chromatograms overlaid on each other quite well and there were no unexplained chemicals in any of them.
Interestingly, I found a piece of literature that states
Also the fact that some instrumental variables have been demonstrated as being non-relevant in the whole range under study can have important practical consequences. For instance the instinctive practice of analysts is to cool the injector as low as possible (−120 ◦ C or even −150 ◦ C) in order to guarantee the inter- mediate trapping of target compounds after twister desorption. This study has shown that this parameter is only relevant down to −60 ◦ C and using lower temperatures represents an unnecessary cost in liquid nitrogen.
K. MacNamara; R. Leardi; F. McGuigana Anal. Chim. Acts., 2009, 636, 190-197. This study was however conducted using the Stir Bar Sorptive Extraction method instead of Tenax, and it doesn't explain the reason for this observation.
Thank you once again!
Sincerely,
Yu Ying