analytic standard for ethylene oxide
Posted: Tue Apr 12, 2016 2:24 pm
by yildizliyim
for ethylene oxide residual test, i need analytic standard for ethylene oxide.
Could you help me about producer company?
Best Regards
Re: analytic standard for ethylene oxide
Posted: Tue Apr 12, 2016 5:48 pm
by GOM
Hi
I found it best to buy pre-prepared standards in solution.
I used to buy it as a standard solution in sealed ampoules from
http://www.greyhoundchrom.com/search.as ... lene+oxide
or
http://www.sigmaaldrich.com/catalog/pro ... ®ion=GB
Originally, I used to make standards by dispensing from a lab cylinder of ethylene oxide but I found that it was far easier and
safer and ultimately cheaper to use the pre-prepared standards
Cool the ampoule in the fridge before opening and further diluting to your required standard.
The analysis was for residual EO in ethoxylated non-ionic detergents
Regards
Ralph
Re: analytic standard for ethylene oxide
Posted: Thu Apr 14, 2016 3:13 pm
by GOM
Hi
As additional comments
1. I used to favour, for economy, buying the 1000ppm standard and then diluting to my required level for standard addition analysis. The diluted standard I would then store in screw cap septum sealed 2ml autosampler vials in the freezer.
2. Ensure that your column/detector can resolve EO from acetaldehyde. which can give a false positive.
Regards
Ralph
Re: analytic standard for ethylene oxide
Posted: Fri Apr 15, 2016 5:35 pm
by James_Ball
Hi
As additional comments
1. I used to favour, for economy, buying the 1000ppm standard and then diluting to my required level for standard addition analysis. The diluted standard I would then store in screw cap septum sealed 2ml autosampler vials in the freezer.
2. Ensure that your column/detector can resolve EO from acetaldehyde. which can give a false positive.
Regards
Ralph
Most of my work has been with volatile analysis, so we use the Mini-inert caps on 1-5ml reaction vials. The slide closure and septum will allow you to draw from the vial repeatedly without losing the volatile analytes. We also store the vials in a freezer as -18C when not in use.
Re: analytic standard for ethylene oxide
Posted: Sat Apr 16, 2016 5:51 pm
by GOM
That is a very useful comment and another way that I hadn't considered in my reply
I have also used these Mini-inert vials successfully
My use of of the 2ml autosampler vials were as a single use disposable standard for economy.. I could also have used 1ml vials.
Regards
Ralph