analytic standard for ethylene oxide
for ethylene oxide residual test, i need analytic standard for ethylene oxide.
Could you help me about producer company?
Hi
I found it best to buy pre-prepared standards in solution.
I used to buy it as a standard solution in sealed ampoules from
http://www.greyhoundchrom.com/search.as ... lene+oxide
or
http://www.sigmaaldrich.com/catalog/pro ... &region=GB
Originally, I used to make standards by dispensing from a lab cylinder of ethylene oxide but I found that it was far easier and safer and ultimately cheaper to use the pre-prepared standards
Cool the ampoule in the fridge before opening and further diluting to your required standard.
The analysis was for residual EO in ethoxylated non-ionic detergents
Ralph
Ralph
Hi
As additional comments
1. I used to favour, for economy, buying the 1000ppm standard and then diluting to my required level for standard addition analysis. The diluted standard I would then store in screw cap septum sealed 2ml autosampler vials in the freezer.
2. Ensure that your column/detector can resolve EO from acetaldehyde. which can give a false positive.
Ralph
Ralph
Hi
Ralph
Most of my work has been with volatile analysis, so we use the Mini-inert caps on 1-5ml reaction vials. The slide closure and septum will allow you to draw from the vial repeatedly without losing the volatile analytes. We also store the vials in a freezer as -18C when not in use.
That is a very useful comment and another way that I hadn't considered in my reply
I have also used these Mini-inert vials successfully
My use of of the 2ml autosampler vials were as a single use disposable standard for economy.. I could also have used 1ml vials.
Ralph
Ralph
In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)