GC 6890N column installation
Posted: Fri Mar 18, 2016 8:03 pm
by uncle-max
Hi everyone,
Iam new to GS, recently I changed column and in the guideline it says to confirm if its correctly installed. I use recommended method but when I press "start" peak appears by itself without injecting sample, and when I inject it nothing happens. What could be the cause? Thank you in advance.
Re: GC 6890N column installation
Posted: Fri Mar 18, 2016 9:09 pm
by dblux_
Hi everyone,
Iam new to GS, recently I changed column and in the guideline it says to confirm if its correctly installed. I use recommended method ...
Let us know what exactly have you done to "confirm".
Posted: Fri Mar 18, 2016 10:01 pm
by uncle-max
Hi everyone,
Thank you for you response.
For inlet:
Mode - split
Heater - 40C
Pressure - 20 psi
Split ratio - 5:1
Oven temp 40C
Detector:
Heater 200C
Reference flow 30
after I injected butane I got a peak I needed but its up side down (it looks down instead of up) how can I change it?
Posted: Fri Mar 18, 2016 10:08 pm
by dblux_
And the type of detector is ...
Is your column capillary or packed ?
Posted: Fri Mar 18, 2016 10:25 pm
by uncle-max
its TCD and its capilary
Posted: Fri Mar 18, 2016 10:31 pm
by dblux_
Is it possible to change the polarity of the signal to get positive peak ?
Posted: Fri Mar 18, 2016 10:49 pm
by uncle-max
Great, thank you so much dblux!
I did it and got a positive peak, I just wondering is it considerate as a right way?
Posted: Fri Mar 18, 2016 10:55 pm
by uncle-max
Another question, what kind of signal plot should I chose? front detector or front column 1?
Posted: Fri Mar 18, 2016 11:43 pm
by dblux_
It is the right way to get positive peaks by chooseing appropriate polarity.
Posted: Fri Mar 18, 2016 11:47 pm
by dblux_
If your detector is installed in front channel then you should choose a signal from front detector.
Posted: Sat Mar 19, 2016 7:58 pm
by uncle-max
Thank you for your answers!!!
The guideline says I need to condition the column and only oven temperature was given, do you know by any change what are other parameters?
Posted: Sat Mar 19, 2016 8:39 pm
by dblux_
Try to "google" column conditioning on manufacturers website. The most important thing is to have carrier gas flow during conditioning. Don't stress the column with too high temperature and too long conditioning.