Hi I've got a fully functional Agilent 1100 system (MWD, autosampler, quaternary pump, degasser, two position column oven). It was run as an MS system but now the MS component is non functional and I want to re configure it to only use the UV detector. I tried doing it from the configuration manager but the system says it requires a new license to run is as an LC instead of an LCMS.

So is there a way I can run methods on an LCMS without the system looking for the MS component? Or am I out of luck?

Thanks ian.
Hello

It is possible. You need to go to tab "instrument" and click "configuration".
Than remove MSD from list and restart Chemstation. MSD will not be visible on system diagram.

Regards

Tomasz Kubowicz
Thanks Tomasz

I did try that but the system still seems to wait for a signal? I think I need to disable the active splitter control (even though that is physically removed)
Hello

In Chemstation you have tab "active spitter". Just turn it off (disable it).

Regards

Tomasz Kubowicz
4 posts Page 1 of 1

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry