HPLC Technique for seperating derivatives of Cholic acid ?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
I'm trying to set up the method which quantitates cholic acid derivatives such as ursodeoxycholic acid, chenodeoxycholic acid, cholic acid, lithocholic acid, deoxycholic acid and hydrodeoxycholic acid at a time.
I'm using Capcellpak C18 column and UV-Detector. As my trial, these substances have very low peak responses at UV-Detector. My question is that is there any way that I can increase the peak sensitivity of them?
I'm using 0.1M phosphate buffer(pH 3.0) and acetonitrile as mobile phase with gradient method. Is my choice is right? Also, is there any way that I can reduce baseline noise ?
It would be really helpful if someone who has a experience dealt with this matter gives me advice.
Thanks for your help in advance.
in my opinion ,you would better check the condition of your uv lamp .because low energy lamp will cut dowm the peak sensitivity of them.
what's more ,you would better clean detector cell .unclean detector cell will rasie baseline noise .make the lod higer than normal . i think if you just want to seperate them .lower flow rate but more polarity mobile phase should be applicable.
looking for laboratory job in USA
in my opinion ,you would better check the condition of your uv lamp .because low energy lamp will cut dowm the peak sensitivity of them.
what's more ,you would better clean detector cell .unclean detector cell will rasie baseline noise .make the lod higer than normal . i think if you just want to seperate them .lower flow rate but more polarity mobile phase should be applicable.
looking for laboratory job in USA

Vanjas

These compounds have very poor UV response. You will need to derivatise them with a UV absorbing compound to improve your detection. I did look at these compounds a long long time ago but I'm afraid I cant recall what the derivatising agent used was. However there are texts which will give you plenty of options. Unfortunately I dont have any to hand.

Regards

Steve

When I used to work at Pickering Laboratories (www.pickeringlabs.com), we sold a custom post-column reaction system for that. Perhaps someone there can find the procedure.

Another alternative is to use evaporative light scattering detection. You will need to switch to a volatile buffer (ammonium acetate or formate).
Mark Tracy
Senior Chemist
Dionex Corp.
5 posts Page 1 of 1

Who is online

In total there are 9 users online :: 2 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo, Semrush [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry