- Posts: 48
- Joined: Fri Sep 24, 2004 8:21 pm
My purpose for this e-mail is because I need any help about a HPLC method development.
I worked with a C18 HPLC platform using a common 5um X 4.6mm X150mm, my question is about the LOD and LOQ.
I tried to quantify (in the same injection) assay and impurities and recently I have problems with the LOD.
My sample preparation was 20mg / 50ml of sample diluent (80/20 methanol/water +0.055 of TFA) and the abundance of the detector was about 2.1Au in a water alliance HPLC (what do you think about this?) and then I want to have my LOD close to 0.01% of the drug substance (assuming that the response factor were the same) and I couldn’t see anything or very tinny peaks (the relation of signal/noise was under 3).
I thought that it was a compromise relation about the saturation of the detector (2.1 AU) and the LOD.What do you think about??? Will be possible to do assay and decomposition in the same injection???.
See the attachments below (chromatograms of LOD and LOQ and sample (25mg/50ml) 100%.
The HPLC parameters are:
Chromatographic Conditions:
Guard Column: None
Analytical Column: Inertsil ODS-3, 150 mm x 4.6 mm, 5-micron particles
Flow Rate: 1.50 mL/minute
Column Temperature: 35° C
Detector Wavelength: 220 nm
Injection Volume: 10 mL
Analysis Time: 25.00 minutes
Equilibration Time: 10 minutes
Needle Wash: 10/90 (v/v) water: methanol with 0.05 % trifluoroacetic acid
Mobile Phase: A: 75/25 (v/v) water/methanol with 0.05% trifluoroacetic acid
B: 20/80 (v/v) water/methanol with 0.05% trifluoroacetic acid
Sample Diluent: 50/50 (v/v) water/methanol with 0.05% trifluoroacetic acid
Gradient Profile
Minutes Curve Mobile Phase
%A %B
0.00 Linear 100.0 0.0
Hold time-Dwell time Linear 100.0 0.0
25.00-Dwell time Linear 0.0 100.0
25.01-Dwell time Linear 100.0 0.0
35.01-Dwell time Linear 100.0 0.0
Thank you in advance
Regards