what is a good way to separate weakly absorbing compounds?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
It may be a naive question, but help me out guys!
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Your question is not clear.
Do you mean "weakly absorbing" such as that you do not see them in the UV detector?
Or do you mean weakly adsorbing as they are not well retained on a reversed-phase column?
If you mean the latter, there are potentially several solutions, such as a C18 that can be used in 100% water, or HILIC, or ion-exchange.
Please describe your problem in a bit more detail!
Do you mean "weakly absorbing" such as that you do not see them in the UV detector?
Or do you mean weakly adsorbing as they are not well retained on a reversed-phase column?
If you mean the latter, there are potentially several solutions, such as a C18 that can be used in 100% water, or HILIC, or ion-exchange.
Please describe your problem in a bit more detail!
-
- Posts: 2
- Joined: Thu Nov 10, 2005 3:32 am
thanks
"weakly absorbing" means "weakly absorbing in UV range"
"weakly absorbing" means "weakly absorbing in UV range"
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
More nitpicking terminology here. The answer to your question is: "You separate them the same way you would separate UV absorbing compounds". The real issue is how to detect them!
The general choices (in order of increasing cost/complexity) are:
- refractive index (RI)
- evaporative light-scattering detectors (ELSD)
- mass spectrometry (LC-MS)
In addition, other techniques may work depending on the specifics of your analyte chemistry:
- conductivity detectior (if your analyte is ionized or ionizable)
- electrochemical detector (if you can easily oxidize or reduce your analyte at an electrode surface)
- post-column reaction detector (if you can react your analyte with something that will make it absorb UV or fluoresce)
- indirect UV detection (if you analyte can displace a UV-absorbing species in the mobile phase)
I'm probably missing some stuff here. If you can provide more details about what you're trying to do, people may be able to give more specific answsers.
The general choices (in order of increasing cost/complexity) are:
- refractive index (RI)
- evaporative light-scattering detectors (ELSD)
- mass spectrometry (LC-MS)
In addition, other techniques may work depending on the specifics of your analyte chemistry:
- conductivity detectior (if your analyte is ionized or ionizable)
- electrochemical detector (if you can easily oxidize or reduce your analyte at an electrode surface)
- post-column reaction detector (if you can react your analyte with something that will make it absorb UV or fluoresce)
- indirect UV detection (if you analyte can displace a UV-absorbing species in the mobile phase)
I'm probably missing some stuff here. If you can provide more details about what you're trying to do, people may be able to give more specific answsers.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 11 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 11 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science