I am currently using the following mobile phases to try and separate an anion compound with HPLC:

A) 0.005M tetrabutyl ammonium phosphate in methanol
B) PIC A reagent in water


With 90% (A) and 10% (B) mixture, pressure seems to be fine and stabilises at around 80 bar at 1ml/min;

But with increase of (B) percentage as little as 20% causes pressure to shoot up as far as 300bar! I found the mixture to be quite viscous as well.

The above was tested without the sample, just the mobile phases washing through the column alone.

Does anyone experienced with similar problems before? Or is this quite abnormal?

Thanks in advance!

Looks like a precipitation of the reagent due to the high methanol content. Disconnect the column, change to water and wash out the precipitated reagentform the instrument. Then reconnect the column and flush it with water at low flow until the pressure goes down.
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry