A method has already been developed in the lab for doing this, but I think previously a different column was being used (reverse phase). There were many problems with this column so a new one was ordered by the other user of the instrument. It is a Waters Spherisorb 5 um Silica column. Not sure if this is ideal for this application, but it's too late now, and I just started using the instrument a couple days ago.
There are already serious problems with the new column. Retention times have quickly reduced and peak resolution is now terrible, after less than a dozen injections. The mobile phase is 70% water/30% acetonitrile. It has been flushed for many hours with water and cleaned once without flipping the column around with a acetonitrile > acetone > hexanes > acetone > acetonitrile gradient. None of this has helped.
Are the analytes possibly adsorbing too strongly in the column? Can anyone suggest an appropriate cleaning or regeneration method? And is this mobile phase appropriate for this type of column? Thanks in advance for any suggestions.