I think that I have it figured out. See, we are running an Agilent GC/MS system with models 6890N and 5973N. On the back of the GC there are a number of different connectors for functions ranging from sampler communications to analog signals outputs. Referring to Agilent document 1215 External Connectors, 6090N, there is only one APG remote connector. Since we are running it with an MS, this connection is already in use. I wasn't quite sure where the cleanest place to wire in would be, nor what type of logic was necessary. The pinouts for the APG remote connector are as follows:
Pin 1: Digital Ground
Pin 2: Prepare
Pin 3: Start
Pin 4: Start Relay
Pin 5: Start Relay
Pin 6: No connection
Pin 7: Ready
Pin 8: Stop
Pin 9: No connection
For remote starting the GC, the logic for prepare, start, and stop inputs are low true, while the ready output is high true.
Agilent also sells a Y-cable for $170.00. This cable should allow me to connect to the MS from the GC and send start/stop signals from a computer to a digital I/O board, to a relay, and then to the GC, which will also remote start the MS. Now all I really need to do, is wait for the part and wire it up when it gets here.
Thanks for the response, Steve. Reading a ready signal and applying a contact closure based on the answer is exactly what I am going to do.