OK, you are asking for help from people on the forum, and you reply to an enquiry for information that will help me help you with a 17 character text string, on which only the first word makes sense.Splitless, 50/0.50
Hi Mark, I am having a very similar problem as the one described in this post. Using a 0.5 uL syringe I'm doing 0.2 uL splitless injections. Samples are volatile fatty acids in 50% methanol 50% water, and same type of column but detection is by FID. My liner also does have wool.0.2 uL splitless? I would definitely start there. Move to a larger volume injection (no, you don't get "backflash") using a split injection; say 1-2 uL with a split ratio of somewhere between 5 and 10.
As MSChemist points out, the small organic acids stick to everything. Using a split injection many times helps that.
I think someone earlier posted that over time more active surfaces accumulate on the liner and septa which the VFAs to stick to and causes carry over.I have the same issue but with GC-FID. The carry over is worse for more concentrated compounds. I currently run 2 blanks between each sample to handle the issue but it would really help to get a more permanent solution.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.