All-

I spend most of my time in the GC realm so I'd appreciate any advice you could give about my HPLC troubles:

Column = J.T. Baker, Weak Cation Exchange (WP CBx), 5 micron, 4.6x250mm
Mobile Phase = 55:20:25 :: Ammonium Chloride, pH 7.7:MeOH:ACN
Flow = 1 mL/min
Sample concentration = 0.2 mg/mL
Inj vol = 20 ul

Analyte = Quaternary amine (curare derived compound)

Problem = severe peak tailing (>2.5) of compound

What type of adjustments can I make to clean this up? Are there any other similar columns (manufacturers, etc.) out there or is this a very specialized phase?

Thanks in advance for any wisdom.
-Jeff
GoldLeader,

First off, I'd suggest lowering the pH. There's no need to use such a high pH with a weak cation exchange material. Lowering the pH will also help minimize any silanol interactions which could be also responsible for the tailing problem.
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry