Tritehyl Citrate Assay
Posted: Thu Oct 20, 2005 11:53 am
Having problems getting the latest USP method for TEC assay to work. Under the current conditions (split ratio 30:1) I am overloading the column and when I increase the split ratio to 100:1 or 150:1 I can't get consistent peak areas.
TEC boiling point is 127C but injector temp only at 85C. Does this seem weird to anyone but me?
Also, can anyone recommend what type of liner to use in the injector? One of my co-workers say I should use the direct inject liner w/out glass wool.
Anyone have any ideas?
Thanks.