Tritehyl Citrate Assay

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Having problems getting the latest USP method for TEC assay to work. Under the current conditions (split ratio 30:1) I am overloading the column and when I increase the split ratio to 100:1 or 150:1 I can't get consistent peak areas.

TEC boiling point is 127C but injector temp only at 85C. Does this seem weird to anyone but me?

Also, can anyone recommend what type of liner to use in the injector? One of my co-workers say I should use the direct inject liner w/out glass wool.

Anyone have any ideas?

Thanks.

That low inlet temperature would seem strange to me as well (does USP call for a temperature-programmable inlet, like in some of their other procedures?). That could be the irreproducible-split reason. My experience has resulted in my disappointment with several USP procedures.
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry