Peak area getting smaller and smaller with every injection!

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hello
I am running US EPA 8270 method on PerkinElmer Clarus580 GC with SQ 8 S Mass Spec. I am extracting for base-neutrals only. I am seeing rapid deterioration of my responses for components and internal standards. Within 15 injections my peak area has almost halved!
I am using deactivated liners - I have tried two different types (PerkinElmer and Restek). I purchase them already deactivated and then dispose of them after use.
I have not known liners to deteriorate so quickly before (I am used to getting at least 40-50 injections out of a liner) so I am wondering if there is something else going on here.
I am kind of at a loss of what else could be the cause. If anyone could help I would really appreciate it.

Thanks!
I don't think that liners chemicaly degrade so quickly - probably its just "mud effect". If you use glass wool - try to minimize its quantity (split mode) or use liner with Jennings cup and increase temp of injector. If you use splitless - try to increase temp of injector and increase split flow.
2 posts Page 1 of 1

Who is online

In total there are 67 users online :: 1 registered, 0 hidden and 66 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 66 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry