A couple of days ago the GC-MS was working just fine. Then I checked on my data files from the overnight run and got this noisy baseline.

https://drive.google.com/file/d/0B7ZnVv ... p=sharing

I did a bake on the column, transfer line and MSD, then tuned and still have this readout. I ran an air and water check and a tune evaluation and both indicated that my nitrogen oxygen ratio were low.

https://drive.google.com/file/d/0B7ZnVv ... sp=sharing

However the air and water check has a large hit for 40, I suspected a contaminated Helium cylinder and switched it but no change occurred. What could be causing this?
Did you get a new delivery of helium recently? Perhaps all of the tanks are contaminated with argon? I had this happen a little while back but when I switched tanks, it went away. Only one tank was contaminated. Perhaps you got a suspicious lot of tanks?
It is a possibility. I had two tanks with water contamination a couple months ago. Thanks! :)
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry