Hi!
I have a serious problem with the GC-MS. It is Thermo Finnigan Trace GC coupled with Polaris GCQ plus.
Last year we upgraded Xcalibur 1.2 to Xcalibur 1.4. Last Friday I tried to open tune window and there has been a message saying that control of the system has been lost would you like to restart? Also the Ion Source Temperature was 0 C.
I pressed ok and everything worked ok and Ion source reached again the temperature set point (200 C).
Last Monday when i opened Tune there was the same message again, but this time when i pressed ok to restart the system there was this message: "System Initialization Error! we had no control at all.
What is worst is that we found the chip that regulates communication between the MS and data system destroyed and the Transfer Line Heater and Temperature Sensor burnt.
The problem is we cannot where the problem starts. Is it the chip that is burnt (and why is this happening?) and then all the other problems follow. Or is the loss of communication between the data system and the ms that creates this or the Transfer Line. The main board is new so as the transfer line.
Thank you!
I think you've permanent problem in your hardware and lost communication then.
It's hard to believe that software fault can do that.

By the way, I'm looking for source assembly of PolarisQ. If some body has it for sale, let me know.
Thank you for answering! Do you mean computer hardware?
No, I mean you mass detector.
If you see a burnt chip, it's probably main problem. but, be aware of other source that can damage your new chip again. better to sent it for repair.
Thanks again!
We have already sent it for repair. I was wondering if someboby had a similar experience with the specific GC-MS in order to find what is causing this so next time we can prevent it. Cheers!
5 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry