Hello all -

Trying to develop a DPPC Content/Assay method using an ELSD on an Acquity H-Class and we have been having issues with repeatability/precision. Basically, recovery of Check Standards and/or Accuracy solutions is very poor, often off by up to 10%.

We are calibrating using a five point log-log linear standard curve and have obtained an R^2 value > 0.995 pretty consistently. Diluent is 75:25:1 MeOH:Chloroform:H20, MP A is 1% Formic in water, MP B is 1% Formic in MeOH. Gradient starts at 10:90, transitions to 2:98, then back to 10:90. We are currently using an Acquity BEH C18 1.7 µm column (We previously tried an Xterra RP8 column with a slightly modified gradient but observed bad peak splitting and decided we were better off with the Acquity). For the ELSD, gas pressure is at 45 psi, nebulizer is at 80% (48C I think), and drift tube is at 40C.

I suspect there is some sort of inherent flaw with the method as we currently have it set up. Maintenance/Calibration was just performed on the detector a few weeks ago. Tailing and plate count look fine, but the appearance of the peak itself is as bad as I've ever seen. I guess I would call it "lumpy". We are using smoothing in our processing method to make it a bit more palatable.

Any suggestions? Is it just something simple that I am completely overlooking?

Thanks.
First of all, did you check solvent miscibility? If I had to shot I would say that solvent from sample wont mix with you mobile phase.

Futhermore I started to work with phosholipids this year and I would consider solvent changing. Dont really remeber but I think that phosphocholine doesnt dissolved in MeOH:CHCl3 2:1 very good. I would go with MeOH:CHCl3 1:2 or n-Hex:IPA 8:2.

You must remember that choline always have a positive charge so interaction with silanols in C18 column will be substiential. I would recomend going to normal phase chromatography - if You want I can give you a nice normal pahse method on Diol column or Silica.
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry