I checked the connections and all of them are just like you said.
And I have to apologize: I couldn't control the Triplus HS because I was adding, at the Chromquest 5.0, the Triplus RSH, that is not the same I use. After a lot of searching in the programm, I realized that I had to add just the Triplus, click with the right buttom on it and then open it. There were an option to select the Triplus HS.
Even after that, when I try to begin a sequence run it doesn't start. The conditions of the GC that were set are OK and the message showed at the GC is: Waiting for the PREP RUN key, and it just stay like that. When I click on button "autosampler" at the GC, it shows me the message saying that there is no autosampler installed. It is strange because when I turn on the Triplus, it makes the normal check of the Autosampler.
I changed the HS turret for the AS turret to see if the message continues, and it continues! I tried to start a sequence run and it started (with AS turret).
I think that the run with HS turret doesn't start because it is heating the syringe and the handshake (at the chromquest I can see a message "Equilibratin Method"). I don't know if is it because I have no access to the autosampler via GC, since it doesn't recognize the autosampler, and not even via chromquest, since it just shows the sets of the equipment.