- Posts: 14
- Joined: Thu Oct 29, 2015 6:48 pm
I am relatively new to RP HPLC, and very new to this forum. In fact, found the forum while searching for answers to my problem. I did find a post close, but not quite the same and from a couple of years ago, so decided might be best to start a fresh post.
I am transferring a tried and true method from a colleague's system to a new system (less than year old). Same C18 column used for both: Zorbax Extend 18, 100 x 2.1mm. 35 minute gradient program is A: methanol; B: water... time table as follows: 0 min 92/8, 15 min 100/0, 20 min 100/0, 21 min 92/8, 35 min 92/8.
On the old system, this was good. On the new system, running the same standard over and over, sequential runs drop in retention time each time from one run to the next. By the end of about 10 runs, the times of a 16 minute peak have dropped down to 10 minutes. Purging the lines with fresh solvents resets all to good, but then repeated runs do the same thing. Switching to different channels gives same result; freshly prepared solvents, same results.
Thanks for any thoughts on this.
Regards,
Mike