Retention time variability according to USP

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello to all,

Does anyone know what is the maximum allowed variability in a compounds retention time according to USP for an HPLC method? This is not clear on the USP general chapter <621>, though some analysts I've read report +/- 10% on the target value.

Or it can only be determined through method development or by means of robustness?

Thank you in advance for your responses :)
I would regard any retention times in USP, a publication, or a technical application as a guideline only. With USP especially, the brand of column is not even detailed, plus there are differences in plumbing among the HPLC instruments.

Now, your retention time REPRODUCIBILITY/REPEATABILITY with your column and your instrument is entirely different. We typically exceed the target of +/- 2% RSD by a large margin.
2 posts Page 1 of 1

Who is online

In total there are 390 users online :: 2 registered, 0 hidden and 388 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 388 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry