3 posts Page 1 of 1
If I have sample that's diluted 1:1 with acetonitrile and my sample is mostly water based I obviously have to take both components into account when calculating potential back flash.

Sorry new to the topic so the question/solution is probably simple for you guys!
Are you trying to estimate the vapour volume from a sample injection, so that you can select the correct inlet liner ? Just calculate for 1/2 ul of acetonitrile and 1/2 ul of water and add them. The GC manufacturers have tools for this on their websites. It is, at best, an inexact science.

Peter
Peter Apps
Plus you will be limited as to the volume you can inject as well as lowering your injection temp. When I had to inject water into an Agilent 6890 using a liner ( internal volume 900ul)
I could only inject 0.5 ul and the injector was just above 100C.
3 posts Page 1 of 1

Who is online

In total there are 83 users online :: 0 registered, 0 hidden and 83 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 83 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry