Tune not passing: mass 50 failing

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hi there!

BFB mass 50 is failing in my tune (13%-14%). I am running 524.2 in Perkin Elmer Clarus 500. Any suggestions about how to troubleshoot?
Increase the ratio of ion 69 to ion 131 and 219, ion 50 is a test of your low end sensitivity.
Increase the ratio of ion 69 to ion 131 and 219, ion 50 is a test of your low end sensitivity.
Exactly.

Also when was the last time the source was cleaned? When running volatiles by purge and trap you don't need to clean often, but you eventually do have to clean it and losing the low masses is one thing that can happen when it gets dirty.
The past is there to guide us into the future, not to dwell in.
Thank you for your answers, guys. My systems just had its annual maintenance so I believe sources are clean.

Following tech support suggestions I set lens 1 to 10 and Lens 2 to 50 along with the LM at 12.5 and Ion Energy Ramp at 1. Problem solved. Hope it will help someone else with a similar issue.
4 posts Page 1 of 1

Who is online

In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry