We have used GC-TCD for hydrogen analysis, and the LOD is about 50-100ppm. I am not sure if there is a detector and/or ways to analyze H2 gas down to 1 ppm ?
Thanks for all input.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Thank you for the help.thohry,
A helium ionization detector will get you down to 1 ppm. Need to be aware that, like the TCD, it will see everything else in your sample. You can look at mine (DBDHID) or Valco's (PDHID) for a retro-fit version that will fit on most GC's. Let me know if you need more info aicmm at flash dot net.
Best regards,
AICMM
Thank you for the offer. but we are not in the UK. Anyway, if in case, I will contact you.We have a Valco one for sale, pre owned, are you in the UK we can install