HP 5890 attenuation issue

Discussions about GC and other "gas phase" separation techniques.

14 posts Page 1 of 1
Old faithful 5890 - this am showed 8333000 as baseline attenuation that can't be changed- happened a year or so ago, was corrected but no one in the lab can remember what corrected it. Anyone ever seen this before?
Are you using Chemstation or an integrator? Does the signal go down if you shut off the flame? Did you do any maintenance or repairs before this happened? The spring from the electrometer PCB to the collector could be shorting or have moisture on it causing it to short. If it's not shorting, could be a bad electrometer.
This might be painfully obvious but I've seen this sort of thing fix stuff in the past. Did you try shutting the system down and bringing it back up?
Hello

Similar value (around 800000) is when there is short circuit in FID.
I'd recommend check FID:
-jet (scratches etc.)
-insulation disks
-spring (perhaps is bent or not aligned properly)

Regards

Tomasz Kubowicz
-spring (perhaps is bent or not aligned properly)
This has happened to me. The spring can get bent and contact the collector.
Are you using Chemstation or an integrator? Does the signal go down if you shut off the flame? Did you do any maintenance or repairs before this happened? The spring from the electrometer PCB to the collector could be shorting or have moisture on it causing it to short. If it's not shorting, could be a bad electrometer.
Thanks, using Chem Station, I believe the "short" must be the case. Will check this out. Thanks
This might be painfully obvious but I've seen this sort of thing fix stuff in the past. Did you try shutting the system down and bringing it back up?
Yes, tried that first- I am checking the collector for a short. Thanks
Hello

Similar value (around 800000) is when there is short circuit in FID.
I'd recommend check FID:
-jet (scratches etc.)
-insulation disks
-spring (perhaps is bent or not aligned properly)

Regards

Tomasz Kubowicz
Thanks, Will check this out.
-spring (perhaps is bent or not aligned properly)
This has happened to me. The spring can get bent and contact the collector.
Thanks, I'll check this out
test
It was the spring- not properly seated from the last overhaul of the FID- there was enough contact to make it work for several months but I could see where the spring was not situated in the "groove" on the collector. Took apart, cleaned,dried and reassembled- doing fine now- however getting the spring properly seated is tough. As you can tell, I'm new to this forum- think I understand the protocols of replying now-Thanks all.
I have a very thin screwdriver that I used to hold that spring off to the side while I slip that collector in there. Then I pull out the screwdriver and the spring snaps into the groove. I'm glad you fixed your problem!
Great idea! Thanks
I have a very thin screwdriver that I used to hold that spring off to the side while I slip that collector in there. Then I pull out the screwdriver and the spring snaps into the groove. I'm glad you fixed your problem!

Do the same thing! Glad your back in business!
14 posts Page 1 of 1

Who is online

In total there are 118 users online :: 2 registered, 0 hidden and 116 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 116 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry