I have speak with AGILENT and they sent me on some track, but the water is always the same level now, but still high enough to give bad result.
We know that it is not the MS, we had isolate our MS and no Water was showed.
The problem is I have more Water than N2, like 2X.
I Know it's not coming from our Helium, we have same line for multiple GC and they are all good.
I have changed all ferrule, septum, o-ring, liners, gold seal, look for leaks with a leak detector found nothing.
I have changed o-ring from EPC too.
I have done a GC Leak check and it shows a variation of 0.2PSI in 10 mins.
I am now trying to see if someone had an idea of what happened to my GC.
Thank you,
S.Chasse