Negative peak with Pulse Flame Photometric Detector

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi all,
I am analyzing Organophosphorus pesticide residue using OI PFPD 5380 version. Recently, I have a problem where there are few negative peaks appear in my chromatogram after I change to a new Column.
There is no leakage any where and I dun know how to solve this problem...any one of you facing this problem before....????

HI along,

with PFPD in general when you have negative peaks it's due to a high concentration of your sample but you sayed that appeared after a change of column............maybe in that case , you have a high bleeding from this column or eventually a high concentration of hydrocarbure in your gas.
With my little experience in PFPD that are the only explanations i can give you for negative peaks........perhaps it could help you!!!!

Thanks BG, what you said are correct, I solve the problem by flushing the column with one end opened and let the detector heated at 350oc for a day.
Thanks
3 posts Page 1 of 1

Who is online

In total there are 424 users online :: 0 registered, 0 hidden and 424 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 424 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry