by
ICKK » Mon Jan 25, 2016 2:29 pm
Yes, you hare right. The NCI ion source give less ion response compared to SCI or (much more less) to EI. This is due mainly to very narrow hole into NCI and SCI sources (locate under filament) compared to EI source. This narrow hole is necessary to keep a relatively high pressure of collision gas into the source only while high vacuum is present in all other parts of vacuum chamber.
Regards NCI and SCI ion sources the first is less capable to produce high abundance of ion due to thermodinamic effects but is highly selective for electron capture chemical species.
At the end, as stated before, to fully use of QP 2010 both EI and NCI sources are necessary. SCI only if you have money to waste.
Regards
Robertino Barcarolo
Thanks a lot Robertino for the quick and informative reply… very much appreciated
I am aware that NCI source is a semi-closed type of ion source and the pressure in the ion source is about 1 to 10 Pa, while CI is a closed type and EI is an open type.
And I do believe that acidic compound and compounds containing electronegative groups give better selectivity and sensitivity using NCI source and that’s what I was trying to get by injecting a carboxylic acid standard. I have the NCI source installed and the standard was checked on its three modes (NCI, SCI and SEI), but the results were reversed as I stated before NCI < SCI < SEI. Would you please send me your email so I can send you some files to explain the output data.
The other thing I would like to check with you please, SCAN vs SIM, again I am getting less peak area applying SIM mode while should be the reverse as I believe. I have read other posts which stated similar problem and they suggested checking the tune file which I did, but no improvement was obtained.
Thanks