HP 5890 attenuation issue
Discussions about GC and other "gas phase" separation techniques.
14 posts
Page 1 of 1
Old faithful 5890 - this am showed 8333000 as baseline attenuation that can't be changed- happened a year or so ago, was corrected but no one in the lab can remember what corrected it. Anyone ever seen this before?
-
- Posts: 196
- Joined: Tue Oct 04, 2011 11:02 pm
Are you using Chemstation or an integrator? Does the signal go down if you shut off the flame? Did you do any maintenance or repairs before this happened? The spring from the electrometer PCB to the collector could be shorting or have moisture on it causing it to short. If it's not shorting, could be a bad electrometer.
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
This might be painfully obvious but I've seen this sort of thing fix stuff in the past. Did you try shutting the system down and bringing it back up?
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
Similar value (around 800000) is when there is short circuit in FID.
I'd recommend check FID:
-jet (scratches etc.)
-insulation disks
-spring (perhaps is bent or not aligned properly)
Regards
Tomasz Kubowicz
Similar value (around 800000) is when there is short circuit in FID.
I'd recommend check FID:
-jet (scratches etc.)
-insulation disks
-spring (perhaps is bent or not aligned properly)
Regards
Tomasz Kubowicz
-
- Posts: 442
- Joined: Mon Jun 13, 2005 7:03 pm
This has happened to me. The spring can get bent and contact the collector.-spring (perhaps is bent or not aligned properly)
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
Thanks, using Chem Station, I believe the "short" must be the case. Will check this out. ThanksAre you using Chemstation or an integrator? Does the signal go down if you shut off the flame? Did you do any maintenance or repairs before this happened? The spring from the electrometer PCB to the collector could be shorting or have moisture on it causing it to short. If it's not shorting, could be a bad electrometer.
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
Yes, tried that first- I am checking the collector for a short. ThanksThis might be painfully obvious but I've seen this sort of thing fix stuff in the past. Did you try shutting the system down and bringing it back up?
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
Thanks, Will check this out.Hello
Similar value (around 800000) is when there is short circuit in FID.
I'd recommend check FID:
-jet (scratches etc.)
-insulation disks
-spring (perhaps is bent or not aligned properly)
Regards
Tomasz Kubowicz
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
Thanks, I'll check this outThis has happened to me. The spring can get bent and contact the collector.-spring (perhaps is bent or not aligned properly)
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
test
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
It was the spring- not properly seated from the last overhaul of the FID- there was enough contact to make it work for several months but I could see where the spring was not situated in the "groove" on the collector. Took apart, cleaned,dried and reassembled- doing fine now- however getting the spring properly seated is tough. As you can tell, I'm new to this forum- think I understand the protocols of replying now-Thanks all.
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
I have a very thin screwdriver that I used to hold that spring off to the side while I slip that collector in there. Then I pull out the screwdriver and the spring snaps into the groove. I'm glad you fixed your problem!
-
- Posts: 8
- Joined: Mon Sep 21, 2015 8:26 pm
Great idea! Thanks
-
- Posts: 196
- Joined: Tue Oct 04, 2011 11:02 pm
I have a very thin screwdriver that I used to hold that spring off to the side while I slip that collector in there. Then I pull out the screwdriver and the spring snaps into the groove. I'm glad you fixed your problem!
Do the same thing! Glad your back in business!
14 posts
Page 1 of 1
Who is online
In total there are 156 users online :: 1 registered, 0 hidden and 155 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Semrush [Bot] and 155 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Semrush [Bot] and 155 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science