by
lmb » Thu Jul 16, 2015 6:57 pm
Argon is expected to produce poorer chromatography compared to helium
Hello Makafui14,
Like the change from any carrier gas to another, the change from helium to argon can produce poorer chromatography. However, nothing in GC theory suggests that argon is "expected" to produce poorer chromatography compared to helium if each gas is used at its own optimal flow rate, or if the relative departure from the optimal flow rate is equal for both gases. So, there is a reason not to give-up right away.
The optimal flow rate of argon is about 3.85 time lower than that of helium. This implies that, in order to preserve the resolutions of all peak pairs when switching from helium to argon, the flow rate of argon should be 3.85 times lower than the flow rate of helium. As a result, the analysis time with argon would be longer than that with helium. How much longer depends on the ratio of the hold-up times with the two gases. That ratio depends not only on the ratio of the flow rates, but also on the difference in viscosities of helium and argon.
In order preserve the resolutions of all peak pairs when switching from one gas to another in temperature-programmed analysis, it is also necessary to re-scale the temperature program in proportion with the ratio of the hold-up times so that the longer is the hold-up time, the proportionally longer is the temperature program.
Summary:
To preserve the resolutions of all peak pairs when switching from helium as a carrier gas to argon in the same column, do the following:
1. Before switching, measure the initial flow rate and the initial hold-up time in the analysis with helium (initial means at initial temperature, 50 C in your case)
2. Choose the initial flow rate of argon 3.85 time lower than the initial flow rate of helium
3. Measure the initial hold-up time in the analysis with argon
4. Calculate the ratio (G) of the measured hold-up times with argon and with helium
5. Rescale the temperature program with argon in proportion with G.
Example. Suppose that G=3 (the hold-up time with argon is 3 times longer than with helium). In this case, instead of the temperature program with helium (50 C (10 mins) at 10 C/min to 120 (5 mins)), use this one: 50 C (30 mins) at 3.33 C/min to 120 (15 mins).
To re-scale the temperature programs, one can also use several GC Method Translation software packages available free of charge from Restek (
http://www.restek.com/ezgc-mtfc), Agilent (
http://www.chem.agilent.com/en-US/Techn ... ation.aspx), and from other sources.
One more thing:
Strictly speaking, the temperature program re-scaling described here is expected to preserve the resolutions of all peak pairs only if both temperature-programmed analyses, with helium and with argon, run in a constant pressure mode. In the constant flow mode, the resolutions of some peak pairs might get better or worse. Usually, the changes in the resolution are very minor. If not then please post your results and we can go from there.
Good luck