- Posts: 6
- Joined: Mon Jul 06, 2015 9:13 am
I'm new to chromatography, so my question might seem stupid. The GC-MS I'm using now had been working well for four years untill recently when we started to observe a lot of siloxanes accumulating in the instrument after some hours of not using it. The performance of the column was gradually worsening and especially it was while analyzing polar compounds. We changed the column and changed the liner and syringe, but it didn't help us, we didn't change a septum, but it is from the same box as the ones used before. From the analysis of the GC-MS data it seems like a septa bleed occurs, becuse m/z 73 coincides with the peaks on the chromotogram and it is dominating in the baseline profile. Also after just several injections of solvents (chloroform and ethyl acetate) with new syringe and liner we observed some dirt in the syringe and inside the liner as well. Moreover, I noticed some metallic particles in the liner and on the surface of the septum. How come this happend? Maybe the syringe touches some metallic parts in the injection port and damages the septum? I appreciate if anyone could help me with the advice on the account, maybe someone encountered something similar.
Thanks!