I think i may need to look at a different column possibly
You may, but that's like saying you may need to look at a new car because your current one is making a funny noise. It's a lot easier (and cheaper) to change something like the aqueous/organic ratio than it is to change the column.
So, first thing: are you getting any retention at all with your current mobile phases? You will need a k' value greater than 2 in order to quantitate effectively (i.e. your retention time should be more than 3x the dead time). You can estimate the dead time (t0) from the column dimensions and the flow rate
t0 ≈ (0.5 * L * dc^2) / F where:
- - L is column length in cm
- dc is column inner diameter in cm
- F is flow in mL/min.
If the peak you are looking at is near that estimated t0, then either you have no retention or your sucralose peak is totally retained. [By the way, sucralose should give a positive peak with RI (see this application note for an example:
http://www2.shimadzu.com/applications/lc/l297.pdf )]. If it is unretained, you could try cutting your organic solvent to 5% (or even zero, as has been suggested). If it's totally retained, you could try increasing the organic solvent concentration (10% increments usually work fairly well - try 25%, 35%, 45% and 55% and see what happens). If you *still* don't see a peak for sucralose, *then* it's time for a different column.