- Posts: 6
- Joined: Thu Apr 09, 2015 8:19 am
The connection is using GPIB, like "PC<GPIB>GC<GPIB>MS"
The problem is happened after power suspension of 36 hours.
When Chemstation is loaded, it can successfully bring up GC, but when in the part of loading MS
Chemstation shows a message "After download, boot in progress was not cleared", and the status is holding in "Initializing"
I can control the parameters of GC like temp, flow..etc.
However, when I go to "Diagnostic and Vacuum control", and check the MS status, it shows “The mass spectrometer interface card is not active”
And the pump of MS is operating normally
I set up a another chemstation without the MS, I can operate my GC-FID normally.....
How can I bring up my MS in chemstation?